Table 1 |
|||||||
|
Summary of studies included in the present evaluation |
|||||||
| Ref. (year) |
Number of patients and category |
Blood glucose target (mmol/l) |
Method |
Used BG meter |
Frequency of measurements |
Hypoglycaemia (mmol/l) |
Results (mmol/l) |
|
|
|||||||
| [1] |
765 Thoracosurgical ICU patients |
4.4–6.1 |
Start at BG >6.1 mmol/l. Insulin ± 0.1–2 IU/hour depending on last two BG values.
Glucose infusion or feeding |
ICU based ABL700® bloodgas/BG analyzer. Arterial blood samples |
1–4 hours |
5.1% of patients <2.2 |
Mean morning BG 5.7 ± 1.1 |
| [4] (2004) |
27 Mixed ICU patients |
4–7 |
Bath IIP: insulin ± 0.5–4 IU/hour, depending on last two BG values |
Accu Check® Advantage 2. Mostly arterial samples |
1–2 hours |
Three BG values <2.2 |
Median BG 6.6 |
| [2] (2004) |
800 Mixed ICU patients |
<7.7 |
Insulin sc. If two BG values >11.1 mmol/l: insulin iv, sliding scale. Glucose infusion/feeding |
Finger stick samples or plasma BG (Vitros® lab analyzer) |
Every 3 hours if sc, hourly if iv |
0.34% of patients <2.2 |
Mean BG 7.2 70% of BG <7.7 |
| [5] (2001) |
20 Critically ill diabetic patients |
6.7 |
Insulin iv between -1.5 and +1.5 IU/hours depending on last two BG values. Glucose-potassium
infusions |
OneTouch® II. Capillary samples |
4 hours |
No BG <2.5 |
Mean BG 7.8 ± 0.2 |
| [6] (2004) |
118 Cardiothoracic ICU patients |
5.5–7.7 |
Yale IIP. Insulin ± 0.5–10 IU, depending on last two BG values and infusion rate |
OneTouch® Surestep Flex |
1 hour |
0.2% of BG <3.3 |
73% between 4.4 and 7.7 |
| [7] (2004) |
52 Medical ICU patients |
5.5–7.7 |
Yale IIP. Insulin ± 0.5–10 IU, depending on last two BG values and infusion rate |
OneTouch® Surestep Flex |
1–4 hours |
0.3% of BG <3.3 |
66% between 4.4 and 7.7 |
| [3] (2004) |
50 Mixed ICU patients |
4.5–6.1 |
Insulin ± 0.5–2 IU/hour, depending on last two BG values. Dextrose infusion or feeding |
Accu Check® Inform. Capillary samples |
1–2 hours |
4% of patients <2.2 |
11.5 ± 3.7 hours/day between 4.5 and 6.1 |
| [8] (2004) |
168 Cardiothoracic ICU patients |
4.4–8.3 |
Insulin 1–16 IU/hour + bolus, sliding scale, depending on last BG value |
Accu Check® Inform. Venous samples |
1–4 hours |
7.1% of BG <2.2 |
61% of BG between 4.4 and 8.3 |
| [9] (2003) |
17 Diabetic patients, acute medical diseases |
6–7 |
GIK + bolus, insulin 1–4 IU/h, dynamic scale, depending on last BG value |
Hemocue® meter. Capillary samples |
1 hour |
Four times |
Mean BG 10.1 |
| [10] (2002) |
37 Postsurgical NIDDM patients |
19 patients sc, 18 patients iv, sliding scale, 5% glucose infusion |
Glucometer®. Capillary samples |
4 hours |
5.6% of patients in iv group |
sc mean: 7.2 ± 1.2 iv mean: 7.3± 1.1 |
|
| [11] (2004) |
72 Cardiothoracic diabetic patients |
6.9–11.1 |
GIK protocol. Continuous GIK infusion + insulin bolus if BG >15 mmol/l |
? |
1 hour |
Mean BG 7.7 ± 0.2 |
|
| [31] (1996) |
60 Surgical NIDDM patients |
3.3–11.1 |
Insulin bolus if BG >11.1 mmol/l. Group 1: saline. Group 2: glucose-insulin. Group
3: bolus every 2 hours |
Capillary samples |
15 min |
No |
Mean BG <9 within all groups |
| [12] (2002) |
29 Diabetic patients, cardiac surgery, 5 days |
6.7–11.1 |
Start if BG >7.8 mol/l. Sliding scale |
Venous and capillary samples |
Six measurements per day |
0.2% of BG <3.8 |
Mean BG 9.5 |
| [13] (1997) |
595 Diabetic patients, cardiac surgery |
<11.1 l |
Portland protocol: insulin depending on last two BG values and insulin infusion rate |
Glucometer |
1–2 hourly |
Mean BG 9.7 |
|
| [14] (1987) |
24 Diabetic patients after surgery |
6.7–10 |
Insulin ± 0.5 IU/hour depending on BG. Bolus if BG >13.3 mol/l. 5% dextrose infusion |
Accu Check + strips |
2 hourly |
1.4% of measurements |
Mean BG between 6.7 and 10 |
| [15] (1988) |
30 Diabetic patients, perioperative |
5–10 |
Group 1: iv, every 4 hours ± 0.5 IU/hour. Group 2: sc, every 4 hours ± 2 IU/4 hours.
Glucose-potassium infusion |
Glucometer |
Hourly during surgery, 4 hourly after surgery |
0.6% of measurements <2.8 in iv group |
67% of iv group between 5.0 and 10; 40% in sc group |
| [17] (2002) |
188 Patients, during cardiac surgery |
4.4–6.6 |
Start (2 IU/hour) with BG >6.6 mmol/l. Double infusion rate until BG <6.6 mol/l. |
? |
20 min |
12% of patients with BG <3.8 |
In 23% of patients BG <8.3 |
| [18] (1994) |
77 Diabetic patients, surgery |
6.7–10 |
Insulin ± 0.5–1.0 U/hr depending on BG |
Reflolux S (+strips) and Glucose hexokinase (lab) |
4 hourly, hourly during surgery |
Two patients |
62% of patients BG between 3.5 and 15.0 |
| [19] (2000) |
24 Type 2 diabetic patients, acute myocardial infarction |
8.3–11.0 |
Insulin ± 1–2 IU/h, depending on BG range. |
Venous samples. Automatic analyzer (lab) |
30 min to 2 hours |
Mean BG 6.9 ± 0.8 |
|
| [20] (2002) |
25 Diabetic patients, acute coronary syndromes |
6.6–8.2 |
Insulin change by -1 to +3 IU/hour, depending on BG |
Beckman® glucose analyzer II |
1–3 hourly |
Four patients with mild hypoglycaemia |
Mean 7.2 ± 1.7 |
| [21] (1999) |
25 Patients, acute stroke (during 24 hours) |
4–7 |
GIK: 500 ml glucose 10% + 16 U insulin + 20 mmol KCl; 100 ml/hour. ± 4 U insulin/infusate,
depending on BG |
BM Glycemic strips |
2 hourly |
One patient |
Mean BG of 68% of patients <7 |
| [22] (1992) |
29 Diabetic patients, acute myocardial infarction |
4–8, to reach within 4 hours |
Sliding scale, more insulin with left ventricular failure and bodyweight >120 % of
ideal |
Capillary samples |
1–4 hourly |
1.2% of BG <3 |
Mean BG 8.2 ± 1.3 |
| [23] (1994) |
158 Diabetic patients, acute myocardial infarction |
7–10 |
>15 mmol/l; bolus iv. Insulin ± 0.5–1 IU/hour depending on BG. Glucose infusion |
Venous samples. Reflolux II |
1–2 hourly |
17.7% of patients with BG <3.0 mmol/l |
Mean BG 9.2 ± 2.9 after 24 hours |
| [24] (1991) |
35 Diabetic patients, acute myocardial infarction |
4–8 |
Sliding scale. Dextrose 5% infusion |
Glucometer II®. Capillary samples |
2–4 hourly |
Mean BG 10.3 ± 2.1 |
|
|
|
|||||||
|
Studies mentioned in the table are arranged according to patient category. Intensive care patients at the top, followed by surgical patients, divided in patients undergoing general surgery and patients undergoing cardiac surgery. The third category of patients consists of patients with an acute myocardial infarction. BG, blood glucose; IIP, insulin infusion protocol; iv, intravenous; NIDDM, noninsulin-dependent diabetes mellitus; sc, subcutaneous. |
|||||||
|
Meijering et al. Critical Care 2006 10:R19 doi:10.1186/cc3981 |
|||||||