Resolution: standard / high Figure 4.
Pressure recordings of proximal airway pressure (Pprox; left panels) and modified
Campbell diagrams (right panels) during simulated spontaneous breathing. Upper panels
show recordings with continuous fresh gas flow; bottom panels show recordings with
the demand-flow system. The left panels depict Pprox variation during two subsequent
breaths. Thin lines represent unfiltered pressure signals and thick lines represent
filtered Pprox. Note the reduced changes in both unfiltered and filtered signal with
the demand-flow system (imposed pressure time product 17 cmH2O s versus 6.8 cmH2O s). Lines in the right panels represent mean lung pressure. Note the reduced surface
area in the lower right panel. Imposed work of breathing is 1.2 J l-1 without the demand-flow system versus 0.5 J l-1 in the lower right panel with the demand-flow system. Pett, pressure at end of the
tracheal tube.
van Heerde et al. Critical Care 2006 10:R103 doi:10.1186/cc4968 |