Abstract
Introduction
Some guidelines to prevent ventilator-associated pneumonia (VAP) do not establish a recommendation for the preferential use of either heat and moisture exchangers (HMEs) or heated humidifiers (HHs), while other guidelines clearly advocate the use of HMEs. The aim of this study was to determine the incidence of VAP associated with HHs or HMEs.
Methods
A randomized study was conducted in the intensive care unit of a university hospital involving patients expected to require mechanical ventilation for >5 days. Patients were assigned to two groups; one group received HH and the other group received HME. Tracheal aspirate samples were obtained on endotracheal intubation, then twice a week, and finally on extubation, in order to diagnose VAP. Throat swabs were taken on admission to the intensive care unit, then twice a week, and finally at discharge from the intensive care unit in order to classify VAP as primary endogenous, secondary endogenous, or exogenous.
Results
A total of 120 patients were assigned to HMEs (60 patients) and HHs (60 patients); 16 patients received mechanical ventilation for less than five days and were excluded from the analysis. Data analysis of the remaining 104 patients (53 HMEs and 51 HHs) showed no significant differences between groups regarding sex, age, Acute Physiology and Chronic Health Evaluation II score, pre-VAP use of antibiotics, days on mechanical ventilation, and diagnosis group. VAP was found in eight of 51 (15.69%) patients in the HH group and in 21 of 53 (39.62%) patients in the HME group (P = 0.006). The median time free of VAP was 20 days (95% confidence interval, 13.34–26.66) for the HH group and was 42 days (95% confidence interval, 35.62–48.37) for the HME group (P <0.001). Cox regression analysis showed the HME as a risk factor for VAP (hazard rate, 16.2; 95% confidence interval, 4.54–58.04; P < 0.001).
Conclusion
The patients mechanically ventilated for more than five days developed a lower incidence of VAP with a HH than with a HME.
Introduction
The use of mechanical ventilation with an artificial airway requires conditioning of the inspired gas [1]. This conditioning is necessary because medicinal gases are cold and dry, and when the upper airway is bypassed it cannot contribute to the natural heat and moisture exchange process of inspired gases.
At low levels of inspired humidity, water is removed from mucous and the periciliary fluid by evaporation, causing increased viscosity of mucous and loss of the periciliary fluid layer. Mucociliary clearance therefore decreases since the thick mucous is difficult for cilia to remove; also, mucociliary transport is impaired due to a decreased cilia beat rate. Continuous desiccation of the mucosa causes cilia paralysis, cell damage, and decreased functional residual capacity, and atelectasis may develop.
Artificial humidification of medicinal gases can be active or passive. In active humidifiers, known as heated humidifiers (HHs), the inspired gas passes across or over a heated water bath. Passive humidifiers, known as artificial noses or heat and moisture exchangers (HMEs), trap heat and humidity from the patient's exhaled gas and return some to the patient on the subsequent inhalation.
Some authors advocate an absolute humidity of 26–30 mg water vapour/l gas [2-5] and have recommended the use of HMEs [6-10]. Other authors advocate an absolute humidity of 44 mg water vapour/l and have recommended the use of HHs [11-14]. There is also controversy concerning the possible influence of these systems on the incidence of ventilator-associated pneumonia (VAP). One study reported a lower incidence of VAP associated with the use of HMEs [15]. On the other hand, several studies found no significant differences in the VAP incidence between the two systems [16-25]. There are also previous data with a lower incidence of VAP associated with HHs [26,27]; one of these studies, however, used an obsolete hydrophobic HME [26], and the other is a congress abstract that was never published [27]. Several published guidelines on the prevention of VAP have not established a recommendation for the preferential use of either HMEs or HHs [28-30], while others clearly advocate the use of HMEs [31,32].
The objective of this study was to compare the incidence of VAP using either a HH system or a HME system in patients on mechanical ventilation for more than five days.
Materials and methods
A randomized study was performed at the 24-bed medicosurgical intensive care unit (ICU) of the Hospital Universitario de Canarias (Tenerife, Spain), a 650-bed tertiary hospital, from 1 January 2005 to 31 December 2005. The study was approved by the Institutional Review Board, and informed consent was obtained from patients or their legal guardians.
We included patients expected to require mechanical ventilation for more than five days. Exclusion criteria were age <18 years, HIV, a blood leukocyte count <1,000 cells/mm3, solid or haematological tumour, and immunosuppressive therapy.
Patients were assigned to receive humidification at the time of intubation either with a HME or with a HH, by a random number list generated using Excel software (Microsoft, Seattle, WA, USA). The HME was the Edith Flex® (Datex-Ohmeda, Helsinki, Finland). HME devices were changed at 48-hour intervals. The HHs used were the MR 850® (Fisher&Paykel Health Care Ltd, Auckland, New Zealand) and the Aerodyne 2000® (Tyco Healthcare/Nellcor, Pleasanton, CA, USA), set to deliver a temperature of 37°C and 100% relative humidity to the proximal airway (containing approximately 44 mg water/l gas as per the manufacturer's recommendations). These HHs are servo-controlled humidifiers with wire-heated circuits without water traps and with an autofeed chamber to refill the chamber with water.
In both patient groups identical measures for the prevention of nosocomial pneumonia were established: no routine change of ventilator circuits, a closed tracheal suction system, a semirecumbent body position, continuous enteric nutrition, periodic verification of the residual gastric volume, prophylactic ranitidine for stress ulcers, oral washing with clorhexidine, no selective digestive decontamination, and no aspiration of subglottic secretions.
Tracheal aspirate samples for analysis were obtained on endotracheal intubation, then twice a week, and finally on extubation, in order to diagnose VAP. Throat swabs were taken on admission to the ICU, then twice a week, and finally at discharge from the unit in order to classify VAP as primary endogenous, secondary endogenous, or exogenous. Necessary clinical samples were taken, entirely independent of how many clinical samples were processed for microbiological analysis.
The diagnosis of pneumonia was established when all of the following criteria were fulfilled: new onset of bronchial purulent sputum, a body temperature >38°C or <35.5°C, a white blood cell count >10,000 mm3 or <4,000/mm3, a chest radiograph showing new or progressive infiltrates, and a significant culture of respiratory secretions by tracheal aspirate (>106 cfu/ml).
Pneumonia was considered as VAP when it was diagnosed after 48 hours of mechanical ventilation. VAP was considered as primary endogenous when caused by microorganisms already present in the patient's oropharyngeal flora on admission to the ICU. VAP was considered as secondary endogenous when caused by microorganisms not found on admission but detected in the patient's oropharyngeal flora during the ICU stay. VAP was considered as exogenous when it was caused by microorganisms that were never carried in the patient's oropharyngeal flora.
The following variables were taken from each patient: sex, age, diagnosis group, Acute Physiology and Chronic Health Evaluation II score, duration of mechanical ventilation, antibiotics previous to VAP, transport out of the ICU, paralytic agents, tracheostomy, reintubation, and mortality.
We found in a previous study [33] that the proportion of patients developing pneumonia after >5 days of mechanical ventilation was 36% using HMEs. For a power of 80% and a 5% type I error, we needed 40 patients per group to test the proportion of patients needed to reduce this proportion (36% using HMEs) to 12% using HHs. We assumed a drop-out rate of 33% (patients <5 days of mechanical ventilation) per group. With this condition, we needed to include 60 patients per group.
Quantitative variables are reported as the mean ± standard deviation, and were compared with the Student t test. Qualitative variables are reported as percentages, and were compared with the chi-squared test or with Fisher's exact test as appropriate. The probability of remaining free of VAP was calculated using the Kaplan-Meier method, and comparison between the two groups was performed with the log-rank test. Five Cox proportional hazard models were constructed for the following dependent variables: VAP-free time, VAP by Gram-positive cocci-free time, VAP by Gram-negative bacilli-free time, endogenous primary VAP-free time, and endogenous secondary VAP-free time. The effect of the Acute Physiology and Chronic Health Evaluation II score was controlled in the five models because the P value was less than 0.20 in the univariate analysis. The main independent variable in the five models was the type of humidification system (HH versus HME). The significant variables were selected using a forward conditional method. P < 0.05 was considered statistically significant. For statistical analyses we used SPSS 12.0.1 software (SPSS Inc., Chicago, IL, USA) and StatXact 5.0.3 software (Cyrus Mehta and Nitin Patel, Cambridge, MA, USA).
Results
A total of 120 patients were randomly assigned to receive HME (n = 60) or to receive HH (n = 60). Of these 120 patients, 16 patients were excluded from the analysis because they received mechanical ventilation for less than five days: in the HME group, six patients were extubated earlier and one patient died; and in the HH group, eight patients were extubated earlier and one patient died.
A total of 104 patients received mechanical ventilation for more than five days (53 patients with HMEs and 51 patients with HHs) and were analyzed. There were no significant differences between groups with respect to sex, age, Acute Physiology and Chronic Health Evaluation II score, use of antibiotics prior to VAP, days on mechanical ventilation, and diagnosis group (Table 1). VAP occurred in eight of 51 (15.69%) patients in the HH group and in 21 of 53 (39.62%) in the HME group (P = 0.006). Kaplan-Meier analysis confirmed a significantly lower incidence of VAP in the HH group than in the HME group (log-rank test = 22.2, P < 0.001) (Figure 1). The median time free of VAP was 20 days (95% confidence interval, 13.34–26.66) for the HH group and was 42 days (95% confidence interval, 35.62–48.37) for the HME group (P <0.001). The multivariate Cox regression analysis showed the HME as a risk factor for VAP (hazard rate, 16.2; 95% confidence interval, 4.54–58.04; P < 0.001), for VAP caused by Gram-positive cocci, for VAP caused by Gram-negative bacilli, for primary endogenous VAP and for secondary endogenous VAP (Table 2).
Figure 1. Cumulative proportion of patients remaining free of ventilator-associated pneumonia
using heat and moisture exchangers (HMEs) versus heated humidifiers (HHs).
Table 1. Characteristics of the heat and moisture exchanger and heated humidifier groups of patients
Table 2. Risk of ventilator-associated pneumonia (VAP) for heat and moisture exchangers versus heated humidifiers, adjusted for the Acute Physiology and Chronic Health Evaluation II score
Table 3 presents the microorganisms responsible for VAP (18 Gram-negative bacilli and 11 Gram-positive cocci) and the pathogenesis of VAP according to oropharyngeal flora (nine primary endogenous, 18 secondary endogenous and two exogenous). In the HME group we found 13 patients with VAP caused by Gram-negative bacilli and eight patients with VAP caused by Gram-positive cocci; the pathogenesis of VAP according to oropharyngeal flora was primary endogenous in eight cases, secondary endogenous in 12 cases, and exogenous in one case. In the HH group we found five patients with VAP caused by Gram-negative bacilli and three patients with VAP caused by Gram-positive cocci; the pathogenesis of VAP according to oropharyngeal flora was primary endogenous in one case, secondary endogenous in six cases, and exogenous in one case.
Table 3. Microorganisms isolated in ventilator-associated pneumonia
Discussion
There are two meta-analyses suggesting an association between the use of HME and a decreased VAP rate [34,35], although only the study of Kirton and colleagues [15] reported a significantly lower incidence of VAP with HMEs compared with HHs. These authors indicate that this effect may be attributed to two mechanisms [15]: the inclusion of a specifically designed microbiological gas filter in the HME, which it is suggested protects the patient from exogenous VAP; and reduced contaminated condensate in the HME circuit [16,19].
In relation to the first mechanism, we think that convincing outcome data are currently insufficient to support the role of gas filtration in reducing the incidence of VAP [36-38]. Previous studies evaluating the effect of gas filtration in anaesthesia machines [36,37] and in ventilators [38] were unable to demonstrate differences in the incidence of VAP between the patient groups with and without filters. In relation to the second mechanism, we agree that the entry of the contaminated condensate circuit into the airway may explain the higher incidence of VAP reported with the HH system. For this reason, in the present study a servo-controlled humidifier was used, which differs from a cascade humidifier in that it has a dual-heated circuit (so the mobile circuit condensate is minimal) and it has an autofeed chamber (eliminating the need to open the circuit to refill the chamber with water), which minimizes the possibility of exogenous microorganisms entering the circuit and causing exogenous VAP.
The reduction of VAP found in our study when using HHs as compared with HMEs may be attributed to three causes. The previously mentioned improvement of the HH system (with a dual-heated circuit and an autofeed chamber) is one such cause. Secondly, the present study analyzed patients on mechanical ventilation for more than five days, and the mean duration of mechanical ventilation (20 days) was higher than in previous studies (4–14 days). Finally, with HHs it is possible to deliver higher levels of humidity to the airway (44 mg water vapour/l gas), and several authors believed that these levels can facilitate maximal mucociliary clearance [14-18].
The results of several studies suggest that humidification is preferable with HHs, reporting a lower incidence of tube occlusion [16,17,26,39], a lower incidence of thick bronchial secretions [16,18,40] and a lower incidence of atelectasis [26] than patients with HMEs. In the study by Hurni and colleagues [41], the state of the ciliated bronchial epithelium, obtained by endotracheal aspirate, was scored on the first day and on day five of mechanical ventilation. In both the HH and the HME groups, patients' scores significantly decreased from day one to day five, and the authors noted that they could not exclude some contribution of an inadequate tracheal temperature and inadequate humidity to this progressive reduction in the cytological score. Also, the reduction of the epithelium score was greater, although not significantly so, in the patient group with the HME system than in the patient group with the HH system. We think that the absence of a significant difference may be for two reasons: the sample size was only 41 patients, and the inspired gas in the HH group was conditioned to a relative humidity of 100% and a temperature of only 32°C (what would the result be with a temperature of 37°C that ensures the delivery of approximately 44 mg water/l gas?).
The lower incidence of VAP found in our study when using HHs as compared with HMEs, both primary endogenous VAP (mostly early onset) and secondary endogenous VAP (mostly late onset), may be attributed to the fact that a HME does not maintain optimal humidification and mucociliary transport beyond 24–48 hours of mechanical ventilation. The review by Williams and colleagues [1], which considered 200 articles/texts on respiratory tract physiology and humidification, reveals that there are few humidity, temperature, and mucosal function studies of human subjects and that the duration in most of them was only 12 hours. The authors proposed that the optimal temperature was that of the body and 100% relative humidity. Exposure times longer than 24 hours need to be studied to fully verify this proposition. The trend in the data of absolute humidity versus exposure time map produced from published series leads us to believe that mucociliary dysfunction can occur later than after 24–48 hours with an absolute humidity level <32 mg/l (the humidification delivered by HMEs).
The present study has several limitations. First, we did not perform a direct assessment of gas heating and humidification in the patient, so the airway temperature and humidity were not monitored (we assumed the reliability of the data reported by the manufacturers). Second, we did not perform indirect assessment of gas heating and humidification in the patient, so we did not examine secretion characteristics or possible epithelial bronchial damage. Finally, the VAP diagnostic procedure was not invasive and we used only tracheal aspirate samples.
Conclusion
Patients mechanically ventilated for more than five days developed a lower incidence of VAP with a HH than with a HME.
Key message
• Patients mechanically ventilated for more than five days developed a lower incidence of VAP with a HH than with a HME.
Abbreivation
HH = heated humidifier; HME = heat and moisture exchanger; ICU = intensive care unit; VAP = ventilator-associated pneumonia.
Competing interests
The authors declare that they have no competing interests.
Authors' contributions
LL was responsible for the conception and design of the study, for data collection, and for analysis and interpretation of the results. ML was responsible for the conception and design of the study, data collection and interpretation of the results. AJ performed statistical analysis and interpretation of the results. MLM and AS were responsible for the conception and design of the study and the interpretation of the results. All authors approved the final version of the manuscript to be published
References
-
Williams R, Rankin N, Smith T, Galler D, Seakins P: Relationship between the humidity and temperature of inspired gas and the function of the airway mucosa.
Crit Care Med 1996, 24:1920-1929. PubMed Abstract | Publisher Full Text
-
Irlbeck D: Normal mechanisms of heat and moisture exchange in the respiratory tract.
Respir Care Clin N Am 1998, 4:189-198. PubMed Abstract
-
McFadden ER Jr, Pichurko BM, Bowman HF, Ingenito E, Burns S, Dowling N, Solway J: Thermal mapping of the airways in humans.
J Appl Physiol 1985, 58:564-570. PubMed Abstract | Publisher Full Text
-
Chatburn R, Primiano F: A rational basis for humidity therapy.
-
Branson RD: Humidification for patients with artificial airways.
-
Jackson C, Webb AR: An evaluation of the heat and moisture exchange performance of four ventilator circuit filters.
Intensive Care Med 1992, 18:264-268. PubMed Abstract | Publisher Full Text
-
Fritz KW, Osterhaus A, Mottner J: The humidification of anesthetic gases during anesthesia using heat and moisture exchangers.
Anaesthesiol Reanim 1992, 17:133-144. PubMed Abstract
-
Kugimiya T, Phuc TG, Numata K: Laboratory evaluation of heat-and-moisture exchangers.
J Anesth 1989, 3:80-85. PubMed Abstract | Publisher Full Text
-
Mebius C: A comparative evaluation of disposable humidifiers.
Acta Anaesthesiol Scand 1983, 27:403-409. PubMed Abstract
-
Sottiaux T, Mignolet G, Damas P, Lamy M: Comparative evaluation of three heat and moisture exchangers during short-term postoperative mechanical ventilation.
Chest 1993, 104:220-224. PubMed Abstract
-
Rankin N: What is optimum humidity?
Respir Care Clin N Am 1998, 4:321-328. PubMed Abstract
-
Luchetti M, Stuani A, Castelli G, Marraro G: Comparison of three different humidification systems during prolonged mechanical ventilation.
Minerva Anestesiol 1998, 64:75-81. PubMed Abstract
-
Lawrence JC: Humidification practices in the Adult Intensive Care Unit, Prince of Wales Hospital.
Respir Care Clin N Am 1998, 4:301-304. PubMed Abstract
-
Ryan SN, Rankin N, Meyer E, Williams R: Energy balance in the intubated human airway is an indicator of optimal gas conditioning.
Crit Care Med 2002, 30:355-361. PubMed Abstract | Publisher Full Text
-
Kirton OC, De Haven B, Morgan J, Morejon O, Civetta J: A prospective, randomized comparison of an in-line heat moisture exchange filter and heated wire humidifiers. Rates of ventilator-associated early-onset (community-acquired) or late-onset (hospital-acquired) pneumonia and incidence of endotracheal tube occlusion.
Chest 1997, 112:1055-1059. PubMed Abstract | Publisher Full Text
-
Martin C, Perrin G, Gevaudan MJ, Saux P, Gouin F: Heat and moisture exchangers and vaporizing humidifiers in the intensive care unit.
Chest 1990, 97:144-149. PubMed Abstract
-
Roustan JP, Kienlen J, Aubas S, Du Cailar J: Comparison of hydrophobic heat and moisture exchangers with heated humidifier during prolonged mechanical ventilation.
Intensive Care Med 1992, 18:97-100. PubMed Abstract | Publisher Full Text
-
Misset B, Escudier B, Rivara D, Leclercq B, Nitenberg G: Heat and moisture exchanger vs heated humidifier during long-term mechanical ventilation. A prospective randomized study.
Chest 1991, 100:160-163. PubMed Abstract
-
Dreyfuss D, Djedaïni K, Gros I, Mier L, Le Bourdellés G, Cohen Y, Estagnasié P, Coste F, Boussougant : Mechanical ventilation with heated humidifiers or heat and moisture exchangers: effects on patient colonization and incidence of nosocomial pneumonia.
Am J Respir Crit Care Med 1995, 151:986-992. PubMed Abstract
-
Kollef MH, Shapiro SD, Boyd V, Silver P, Von Harz B, Trovillion E, Prentice D: A randomized clinical trial comparing an extended-use hygroscopic condenser humidifier with heated-water humidification in mechanically ventilated patients.
Chest 1998, 113:759-767. PubMed Abstract | Publisher Full Text
-
Branson R, Davis K, Campbell RS, Johnson DJ, Porembka DT: Humidification in the intensive care unit. Prospective study of a new protocol utilizing heated humidification and a hygroscopic condenser humidifier.
Chest 1993, 104:1800-1805. PubMed Abstract
-
Boots RJ, Howe S, George N, Harris FM, Faoagali J: Clinical utility of hygroscopic heat and moisture exchangers in intensive care patients.
Crit Care Med 1997, 25:1707-1712. PubMed Abstract | Publisher Full Text
-
Memish ZA, Oni GA, Djazmati W, Cunningham G, Mah MW: A randomized trial to compare the effects of a heat and moisture exchanger with a heated humidifying system on the occurrence rate of ventilator-associated pneumonia.
Am J Infect Control 2001, 29:301-305. PubMed Abstract | Publisher Full Text
-
Lacherade JC, Auburtin M, Cerf C, Van de Louw A, Soufir L, Rebufat Y, Rezaiguia S, Ricard JD, Lellouche F, Brun-Buisson C, Brochard L: Impact of humidification systems on ventilator-associated pneumonia: a randomized multicenter trial.
Am J Respir Crit Care Med 2005, 172:1276-1282. PubMed Abstract | Publisher Full Text
-
Kranabetter R, Leier M, Kammermeier D, Just HM, Heuser D: The effects of active and passive humidification on ventilation-associated nosocomial pneumonia.
Anaesthesist 2004, 53:29-35. PubMed Abstract | Publisher Full Text
-
Cohen IL, Weinberg PF, Alan I, Rowinsky GS: Endotracheal tube occlusion associated with the use of heat and moisture exchangers in the intensive care unit.
Crit Care Med 1988, 16:277-279. PubMed Abstract
-
Blin F, Fraïsse F, Chauveau P, Lahilaire P, Boulet E, Bouhaddi R, Sadourny F, Schlatter J, Trouillet G, Hermitte M, Kleinknecht D, the members of the Nosocomial Infections Research Group: Incidence of nosocomial pneumonias among 1788 ventilated patients in 6 ICU according to the type of humidification used.
Intensive Care Med 1996, 22(Suppl 3):S324.
[abstract]
-
Torres A, Carlet J, the European Task Force on Ventilator Associated Pneumonia: Ventilator associated pneumonia.
Eur Respir J 2001, 17:1034-1045. PubMed Abstract | Publisher Full Text
-
American Thoracic Society and Infectious Diseases Society of America: Guidelines for the management of adults with hospital-acquired, ventilator associated, and healthcare associated pneumonia.
Am J Respir Crit Care Med 2005, 171:388-416. PubMed Abstract | Publisher Full Text
-
Tablan OC, Anderson LJ, Besser R, Bridges C, Hajjeh R, CDC and Healthcare Infection Control Practices Advisory Committee: Guidelines for preventing health-care – associated pneumonia, 2003: recommendations of CDC and the Healthcare Infection Control Practices Advisory Committee.
MMWR Recomm Rep 2004, 53:1-36. PubMed Abstract | Publisher Full Text
-
Dodek P, Keenan S, Cook D, Heyland D, Jacka M, Hand L, Muscedere J, Foster D, Mehta N, Hall R, Brun-Buisson C, the Canadian Critical Care Trials Group, Canadian Critical Care Society: Evidence based clinical practice guideline for the prevention of ventilator associated pneumonia.
Ann Intern Med 2004, 141:305-313. PubMed Abstract
-
Hess DR, Kallstrom TJ, Mottram CD, Myers TR, Sorenson HM, Vines DL: American Association for Respiratory Care. Care of the ventilator circuit and its relation to Guidelines for ventilator associated pneumonia.
Respir Care 2003, 48:869-879. PubMed Abstract
-
Lorente L, Lecuona M, Jiménez A, Mora ML, Sierra A: Tracheal suction by closed system without daily change versus open system.
Intensive Care Med 2006, 32:538-544. PubMed Abstract | Publisher Full Text
-
Cook D, De Jonghe B, Brochard L, Brun-Buisson C: Influence of airway management on ventilator-associated pneumonia: evidence from randomized trials.
JAMA 1998, 279:781-787. PubMed Abstract | Publisher Full Text
-
Kola A, Eckmanns T, Gastmeier P: Efficacy of heat and moisture exchangers in preventing ventilator-associated pneumonia: meta-analysis of randomized controlled trials.
Intensive Care Med 2005, 31:5-11. PubMed Abstract | Publisher Full Text
-
Garibaldi RA, Britt MR, Webster C, Pace NL: Failure of bacterial filters to reduce the incidence of pneumonia after inhalation anesthesia.
Anesthesiology 1981, 54:364-368. PubMed Abstract
-
Feeley TW, Hamilton WK, Xavier B, Moyers J, Eger EI: Sterile anesthesia breathing circuits do not prevent postoperative pulmonary infection.
Anesthesiology 1981, 54:369-372. PubMed Abstract
-
Lorente L, Lecuona M, Málaga J, Revert C, Mora ML, Sierra A: Bacterial filters in respiratory circuits: an unnecessary cost?
Crit Care Med 2003, 31:2126-2130. PubMed Abstract | Publisher Full Text
-
Hess DR: And now for the rest of the store.
Respir Care 2002, 47:696-699. PubMed Abstract
-
Luchetti M, Stuani A, Castelli G, Marraro G: Comparison of three different humidification systems during prolonged mechanical ventilation.
Minerva Anestesiol 1998, 64:75-81. PubMed Abstract
-
Hurni JM, Feihl F, Lazor R, Leuenberger P, Perret C: Safety of combined heat and moisture exchanger filters in long-term mechanical ventilation.
Chest 1997, 111:686-691. PubMed Abstract | Publisher Full Text





