Dual-lumen catheters for continuous venovenous hemofiltration: limits for blood delivery via femoral vein access and a potential alternative in an experimental setting in anesthetized pigs

Juliane K Unger1 email, Klaus Pietzner1 email, Roland C Francis2 email, Juergen Birnbaum3 email, Marc Michael Theisen4 email, Arne-Joern Lemke5 email and Stefan M Niehues5 email

1Department of Comparative Medicine and Laboratory Animal Sciences, Charité Campus Virchow-Klinikum, Universitätsmedizin Berlin, Augustenburger Platz 1, D-13353 Berlin, Germany

2Department of Anesthesiology and Intensive Care Medicine, Charité Campus Virchow-Klinikum, Universitätsmedizin Berlin, Augustenburger Platz 1, D-13353 Berlin, Germany

3Department of Anesthesiology and Intensive Care Medicine, Charité Campus Mitte, Universitätsmedizin Berlin, Charitéplatz 1, D-10117 Berlin, Germany

4Department of Anesthesiology and Intensive Care, University Hospital, Albert-Schweitzer-Str. 33, D-48149 Muenster, Germany

5Department of Radiology, Charité Campus Virchow-Klinikum, Universitätsmedizin Berlin, Augustenburger Platz 1, D-13353 Berlin, Germany

author email corresponding author email

Critical Care 2007, 11:R18doi:10.1186/cc5693

Published: 15 February 2007

Abstract

Introduction

Small intravascular volume, pathophysiological hemorheology, and/or low cardiac output [CO] are assumed to reduce available blood flow rates via common dual-lumen catheters (except for those with a right atrium catheter tip position) in the critically ill patient. We performed an experimental animal study to verify these assumptions.

Methods

Anesthetized, ventilated pigs (35 to 50 kg) were allocated to different hemorheological conditions based on the application of different volume substitutes (that is, colloids and crystalloids, n = 6 to 7 per volume substitute). In a second step, allocation to the final study group was performed after the determination of the highest values for access flow (Qa) via an axial dual-lumen catheter (11 French, 20 cm long, side holes) placed in the femoral vein. High Qa rates (>300 ml/minute) were allocated to the dual-lumen catheter group; low Qa rates were switched to a 'dual-vein approach' using an alternative catheter (8.5-French sheath) for separate blood delivery. Hemodynamics (CO and central venous pressure [CVP]) and blood composition (blood cell counts, plasma proteins, and colloid osmotic pressure) were measured. Catheter tip positions and vessel diameters were exemplified by computed tomography.

Results

Forty-four percent of the animals required an alternative vascular access due to only minimal Qa via the dual-lumen catheter. Neither hemorheologically relevant aspects nor CO and CVP correlated with the Qa achievable via the femoral vein access. Even though the catheter tip of the alternative catheter provided common iliac vein but not caval vein access, this catheter type enabled higher Qa than the dual-lumen catheter positioned in the caval vein.

Conclusion

With respect to the femoral vein approach, none of the commonly assumed reasons for limited Qa via the arterial line of an axial dual-lumen catheter could be confirmed. The 8.5-French sheath, though not engineered for that purpose, performed quite well as an alternative catheter. Thus, in patients lacking right jugular vein access with tip positioning of large-French dual-lumen catheters in the right atrium, it would be of interest to obtain clinical data re-evaluating the 'dual-vein approach' with separate blood delivery via a tip-hole catheter in order to provide high-volume hemofiltration.