Resolution: standard / high Figure 5.
Comparison of end expiratory lung volume (EELV) measured by the helium dilution technique
and the nitrogen washout/washin method. (a) The EELV measured by the helium dilution as a function of the EELV measured
by nitrogen washout/washin method (EELV helium dilution = (290 + 0.92) × EELV GE,
r2 = 0.82, p < 0.00001). (b) The Bland-Altman plot of the EELV measured with the nitrogen
washout/washin technique and the EELV measured with the helium dilution method. The
x axis shows the mean of the two measurements and the y axis shows the difference
between the EELV measured by then helium dilution method and the nitrogen washoutwashin
measured by the CT scan (average difference -229 ± 164 ml, limits of agreement -558
– 100 ml).
Chiumello et al. Critical Care 2008 12:R150 doi:10.1186/cc7139 |