Critical Care Volume 13 Issue 1 |
Viewing options:Associated material:Related literature:- Articles citing this article
- Other articles by authors
- Related articles/pages
Tools: Post to:
|
ReviewBench-to-bedside review: The importance of the precision of the reference technique in method comparison studies – with specific reference to the measurement of cardiac outputMaurizio Cecconi1,2 , Andrew Rhodes2 , Jan Poloniecki3 , Giorgio Della Rocca1 and R Michael Grounds2  1Department of Anesthesia and Intensive Care, Azienda Ospedaliero Universitaria Udine, Piazzale Santa Maria della Misericordia, 33100 Udine, Italy 2Department of Intensive Care Medicine, St George's Hospital, London, SW17 0QT, UK 3Community Health Sciences, St George's, University of London, SW17 0RE, UK author email corresponding author email
Critical Care 2009,
13:201doi:10.1186/cc7129
|
|
| Published: |
13 January 2009 |
Abstract
Bland-Altman analysis is used for assessing agreement between two measurements of the same clinical variable. In the field of cardiac output monitoring, its results, in terms of bias and limits of agreement, are often difficult to interpret, leading clinicians to use a cutoff of 30% in the percentage error in order to decide whether a new technique may be considered a good alternative. This percentage error of ± 30% arises from the assumption that the commonly used reference technique, intermittent thermodilution, has a precision of ± 20% or less. The combination of two precisions of ± 20% equates to a total error of ± 28.3%, which is commonly rounded up to ± 30%. Thus, finding a percentage error of less than ± 30% should equate to the new tested technique having an error similar to the reference, which therefore should be acceptable. In a worked example in this paper, we discuss the limitations of this approach, in particular in regard to the situation in which the reference technique may be either more or less precise than would normally be expected. This can lead to inappropriate conclusions being drawn from data acquired in validation studies of new monitoring technologies. We conclude that it is not acceptable to present comparison studies quoting percentage error as an acceptability criteria without reporting the precision of the reference technique. |