Self-calibrating pulse contour cardiac output: do validation studies really show its clinical reliability?

Lester AH Critchley

Author Affiliations

Department of Anaesthesia and Intensive Care, The Chinese University of Hong Kong, Prince of Wales Hospital, Shatin, New Territories, Hong Kong, PR China

Critical Care 2009, 13:123 doi:10.1186/cc7718


See related research by Senn et al., http://ccforum.com/content/13/2/R32

Published: 4 March 2009

Abstract

The present study was performed to test a new software version of the FloTrac/Vigileo using head-up–head-down tilting in post-cardiac surgery patients. Impressive improvements in Bland and Altman limits of agreement from 37.5% to 21.6% were recorded. The results, however could be attributed to a failure to produce a wide enough range of test circulatory conditions. A more rigorous test of performance is needed before any real conclusion concerning use of the FloTrac/Vigileo in clinical practice can be made.