Assessment of cardiac output changes using a modified FloTrac/Vigileo™ algorithm in cardiac surgery patients

Alban Senn, Danny Button, Andreas Zollinger and Christoph K Hofer*

Author Affiliations

Institute of Anaesthesiology and Intensive Care Medicine, Triemli City Hospital Zurich, Birmensdorferstrasse 497, 8063 Zurich, Switzerland

For all author emails, please log on.

Critical Care 2009, 13:R32 doi:10.1186/cc7739

Published: 4 March 2009

Abstract

Introduction

The FloTrac/Vigileo™ (Edwards Lifesciences, Irvine, CA, USA) allows pulse pressure-derived cardiac output measurement without external calibration. Software modifications were performed in order to eliminate initially observed deficits. The aim of this study was to assess changes in cardiac output determined by the FloTrac/Vigileo™ system (FCO) with an initially released (FCOA) and a modified (FCOB) software version, as well as changes in cardiac output from the PiCCOplus™ system (PCO; Pulsion Medical Systems, Munich, Germany). Both devices were compared with cardiac output measured by intermittent thermodilution (ICO).

Methods

Cardiac output measurements were performed in patients after elective cardiac surgery. Two sets of data (A and B) were obtained using FCOA and FCOB in 50 patients. After calibration of the PiCCOplus™ system, triplicate FCO and PCO values were recorded and ICO was determined in the supine position and cardiac output changes due to body positioning were recorded 15 minutes later (30° head-up, 30° head-down, supine). Student's t test, analysis of variance and Bland-Altman analysis were calculated.

Results

Significant changes of FCO, PCO and ICO induced by body positioning were observed in both data sets. For set A, ΔFCOA was significantly larger than ΔICO induced by positioning the head down. For set B, there were no significant differences between ΔFCOB and ΔICO. For set A, increased limits of agreement were found for FCOA-ICO when compared with PCO-ICO. For set B, mean bias and limits of agreement were comparable for FCOB-ICO and PCO-ICO.

Conclusions

The modification of the FloTrac/Vigileo™ system resulted in an improved performance in order to reliably assess cardiac output and track the related changes in patients after cardiac surgery.