ResearchElectrical impedance tomography compared to positron emission tomography for the measurement of regional lung ventilation: an experimental studyJC Richard1,2,3 , C Pouzot2,4 , A Gros1 , C Tourevieille5 , D Lebars5 , F Lavenne5 , I Frerichs6 and C Guérin1,2,3  1Service de Réanimation Médicale et d'Assistance Respiratoire, Hôpital de la Croix Rousse 103 Grande Rue de la Croix Rousse, Lyon, 69004, France 2Creatis, Centre National de la Recherche Scientifique Unité Mixte de Recherche 5220 and Institut National de la Santé et de l'Enseignement et de la Recherche Médicale U 630, 7 avenue Jean Capelle, Villeurbanne, 69621 Cedex, France 3Université de Lyon, Université Claude Bernard Lyon 1, 8 avenue Rockefeller, Lyon, 69008, France 4Service de Soins Intensifs Animaux et Medecine d'Urgence, Ecole Nationale Vétérinaire de Lyon, 1 Avenue Bourgelat, Marcy L'Etoile, 69280, France 5Centre de Recherche Médicale par Emission de Positrons, Imagerie du vivant, 59 Boulevard Pinel, 69003, Lyon, France 6Anaesthesiology and Intensive Care Medicine, University Medical Centre Schleswig-Holstein, Kiel, Germany author email corresponding author email
Critical Care 2009,
13:R82doi:10.1186/cc7900 Abstract
Introduction
Electrical impedance tomography (EIT), which can assess regional lung ventilation at the bedside, has never been compared with positron-emission tomography (PET), a gold-standard to quantify regional ventilation. This experiment systematically compared both techniques in injured and non-injured lungs.
Methods
The study was performed in six mechanically ventilated female piglets. In normal lungs, tidal volume (VT) was randomly changed to 6, 8, 10 and 15 ml/kg on zero end-expiratory pressure (ZEEP), then, at VT 10 ml/kg, positive end-expiratory pressure (PEEP) was randomly changed to 5, 10 and 15 cmH2O. Afterwards, acute lung injury (ALI) was subsequently created in three animals by injecting 3 ml/kg hydrochloric acid into the trachea. Then at PEEP 5 cmH2O, VT was randomly changed to 8 and 12 ml/kg and PEEP of 10 and 15 cmH2O applied at VT 10 ml/kg. EIT and PET examinations were performed simultaneously. EIT ventilation (VTEIT) and lung volume (VL) were measured in the anterior and posterior area of each lung. On the same regions of interest, ventilation (VPET) and aerated lung volume (VAatten) were determined with PET.
Results
On ZEEP, VTEIT and VPET significantly correlated for global (VTEIT = VPET - 2E-13, R2 = 0.95, P < 0.001) and regional (VTEIT = 0.81VPET+7.65, R2 = 0.63, P < 0.001) ventilation over both conditions. For ALI condition, corresponding R2 were 0.91 and 0.73 (P < 0.01). Bias was = 0 and limits of agreement were -37.42 and +37.42 ml/min for global ventilation over both conditions. These values were 0.04 and -29.01 and +29.08 ml/min, respectively, for regional ventilation. Significant correlations were also found between VL and VAatten for global (VL = VAatten+1E-12, R2 = 0.93, P < 0.0001) and regional (VL = 0.99VAatten+0.92, R2 = 0.65, P < 0.001) volume. For ALI condition, corresponding R2 were 0.94 (P < 0.001) and 0.54 (P < 0.05). Bias was = 0 and limits of agreement ranged -38.16 and +38.16 ml for global ventilation over both conditions. These values were -0.24 and -31.96 to +31.48 ml, respectively, for regional ventilation.
Conclusions
Regional lung ventilation and volume were accurately measured with EIT in healthy and injured lungs and validated by simultaneous PET imaging. |