Table 1 |
|||
|
Inter technology agreements based on 20 values for each technology |
|||
|
NICOM vs. PICCO TD |
PICCO PC vs. PICCO TD |
NICOM vs. PICCO PC |
|
|
|
|||
|
Bias |
0,26 |
0,04 |
0,22 |
|
R |
0,76 |
0,77 |
0,55 |
|
RE |
0,12 |
0,11 |
0,18 |
|
LOA |
1,52 |
1,77 |
2,41 |
|
CV |
0,31 |
0,38 |
0,49 |
|
C&C |
0,93 |
0,93 |
0,80 |
|
|
|||
|
Values of cardiac output obtained during positive end expiratory pressure were excluded from this analysis to minimize the impact of differences in time responsiveness between the devices. R = coefficient of determination for inter-technologies comparisons; RE = averaged value of relative error (absolute bias/mean); LOA = limits of agreements of the Bland and Altman representation (two × standard deviations of the difference) in L/min; CV = coefficient of variation (two × standard deviation/mean); C&C = percentage of the differences inside the 30% limit of acceptability as suggested by Critchley and Critchley [16]. |
|||
|