This article is part of the supplement: 29th International Symposium on Intensive Care and Emergency Medicine
Adaptive support ventilation prevents ventilator-induced diaphragmatic dysfunction: an in vivo piglet study
1 Saint-Eloi Hospital, Montpellier, France
2 Hamilton Medical, Rhazuns, Switzerland
Critical Care 2009, 13(Suppl 1):P31 doi:10.1186/cc7195
The electronic version of this article is the complete one and can be found online at: http://ccforum.com/content/13/S1/P31
| Published: | 13 March 2009 |
© 2009 Jung et al; licensee BioMed Central Ltd.
Introduction
Mechanical ventilation is a lifesaving supportive therapy for patients with acute respiratory failure. However, prolonged mechanical ventilation results in the complete absence of neural activation and mechanical activity of the diaphragm and has been shown to induce ventilator-induced diaphragmatic dysfunction (VIDD) [1]. Few studies have shown that maintaining spontaneous ventilation could prevent VIDD in in vitro animal studies [2]. Adaptive support ventilation (ASV) is an automatic ventilation mode that allowed pressure-controlled breaths in active patients able to trigger. The aim of our study was to compare ASV with controlled ventilation (CV) without paralysing agents on diaphragmatic contractile properties in an in vivo piglet study.
Methods
Two groups of six anesthetized piglets were ventilated during a 72-hour period. Piglets in the CV group (n = 6) were ventilated without any spontaneous ventilation and piglets in the ASV group (n = 6) were ventilated with the possibility of triggering spontaneous ventilation. The main endpoint was the transdiaphragmatic pressure (Pdi) (after bilateral, supramaximal, transjugular stimulation of the two phrenic nerves), which represents the in vivo contractile function of the diaphragm. A force-frequency curve was drawn after stimulation from 20 to 120 Hz of the phrenic nerves.
Results
The piglets in the ASV group were maintained with spontaneous ventilation during 80% of the study period instead of less than 1% in the CV group. At 72 hours, Pdi was decreased by 35% in the CV group although it was not modified in the ASV group (Figure 1).
Figure 1. Pdi evolution during the study. P < 0.05 CV vs. ASV group at 48 hours.
Conclusion
Spontaneous breathing with ASV prevents VIDD in comparison with totally CV in an in vivo healthy piglet model.
References
-
Sassoon CS: Ventilator-associated diaphragmatic dysfunction.
Am J Respir Crit Care Med 2002, 166:1017-1018. PubMed Abstract | Publisher Full Text
-
Futier E, Constantin JM, Combaret L, Mosoni L, Roszyk L, Sapin V, et al.: Pressure support ventilation attenuates ventilator-induced protein modifications in the diaphragm.
Crit Care 2008, 12:R116. PubMed Abstract | BioMed Central Full Text | PubMed Central Full Text