Table 1 |
|||
|
In vitro dynamic test of cytokine adsorption by Amberchrom CG161 M resin |
|||
|
Cytokine |
Concentration (pg/ml) |
Amount bound (% leakage) (pg/ml adsorbent) |
Theoretical binding capacity (pg/ml adsorbent) |
|
|
|||
|
TNF-α |
675 ± 72 |
4999 (5%) ± 1703 |
200,980 ± 53,607 |
|
IL-1-β |
331 ± 1 |
61,439 (5%) ± 10,459 |
116,842 ± 24,707 |
|
IL-6 |
334 ± 59 |
35,729 (5%) ± 8149 |
912,776 ± 200,096 |
|
IL-8 |
340 ± 186 |
322 (12%) ± 161 |
48,940 ± 26,089 |
|
IL-10 |
464 ± 143 |
69,187 (5%) ± 31,112 |
972,880 ± 61,006 |
|
|
|||
|
The conditions used for the dynamic studies on Amberchrom CG161 M resin are extensively described in the Materials and Methods. Briefly, the resin columns were perfused at a low linear velocity (0.30 ml/min). The theoretical binding capacity of the resins for individual cytokines which was calculated by extrapolation of the increase in cytokine concentration at the column outlet over time only gives a rough estimation of the actual maximal binding capacity. |
|||
|
Cantaluppi et al. Critical Care 2010 14:R4 doi:10.1186/cc8835 |
|||