Accuracy and precision of end-expiratory lung-volume measurements by automated nitrogen washout/washin technique in patients with acute respiratory distress syndrome
-
* Corresponding author: Jean Dellamonica jean.dellamonica@inserm.fr
1 Réanimation Médicale, AP-HP, Centre Hospitalier Albert Chenevier, Henri Mondor, avenue Marechal de Lattre de Tassigny, Créteil, 94000, France
2 Réanimation Médicale, CHU de Nice, Hôpital L'Archet, Université de Nice Sophia Antipolis, Route de St Antoine de Ginestière, Nice 06200, France
3 Réanimation Médicale, AP-HP, Hôpital Européen Georges Pompidou, rue Leblanc, Paris 75015, France
4 Réanimation Médicale, CHU Angers, rue Larrey, Angers 49100, France
5 Réanimation Médicale & UPRES EA 3830, CHU Charles Nicolle, rue Germont, Rouen 76031, France
6 Pneumologia Ospedale San Paolo, Universita degli Studi di Milano, via A. di Rudini 8, Milano 20142, Italy
7 Servei de Medicina Intensiva, Hospital de Sant Pau, C. Sant Quinti 89, Barcelona, 08041, Spain
8 Réanimation Polyvalente, AP-HP, Hôpital Pitié Salpêtrière, UPMC, Université Paris 6, Boulevard de l'Hôpital, Paris 75014, France
9 INSERM U-955, Université Paris EST, avenue Marechal de Lattre de Tassigny, Créteil 94000, France
10 Intensive Care Department, University Hospital and University of Geneva, rue Gabrielle Perret-Gentil, Geneva 1211, Switzerland
Critical Care 2011, 15:R294 doi:10.1186/cc10587
The electronic version of this article is the complete one and can be found online at: http://ccforum.com/content/15/6/R294
| Received: | 29 July 2011 |
| Revisions received: | 4 December 2011 |
| Accepted: | 7 December 2011 |
| Published: | 7 December 2011 |
© 2011 Dellamonica et al.; licensee BioMed Central Ltd.
This is an open access article distributed under the terms of the Creative Commons Attribution License (http://creativecommons.org/licenses/by/2.0), which permits unrestricted use, distribution, and reproduction in any medium, provided the original work is properly cited.
Abstract
Introduction
End-expiratory lung volume (EELV) is decreased in acute respiratory distress syndrome (ARDS), and bedside EELV measurement may help to set positive end-expiratory pressure (PEEP). Nitrogen washout/washin for EELV measurement is available at the bedside, but assessments of accuracy and precision in real-life conditions are scant. Our purpose was to (a) assess EELV measurement precision in ARDS patients at two PEEP levels (three pairs of measurements), and (b) compare the changes (Δ) induced by PEEP for total EELV with the PEEP-induced changes in lung volume above functional residual capacity measured with passive spirometry (ΔPEEP-volume). The minimal predicted increase in lung volume was calculated from compliance at low PEEP and ΔPEEP to ensure the validity of lung-volume changes.
Methods
Thirty-four patients with ARDS were prospectively included in five university-hospital intensive care units. ΔEELV and ΔPEEP volumes were compared between 6 and 15 cm H2O of PEEP.
Results
After exclusion of three patients, variability of the nitrogen technique was less than 4%, and the largest difference between measurements was 81 ± 64 ml. ΔEELV and ΔPEEP-volume were only weakly correlated (r2 = 0.47); 95% confidence interval limits, -414 to 608 ml). In four patients with the highest PEEP (≥ 16 cm H2O), ΔEELV was lower than the minimal predicted increase in lung volume, suggesting flawed measurements, possibly due to leaks. Excluding those from the analysis markedly strengthened the correlation between ΔEELV and ΔPEEP volume (r2 = 0.80).
Conclusions
In most patients, the EELV technique has good reproducibility and accuracy, even at high PEEP. At high pressures, its accuracy may be limited in case of leaks. The minimal predicted increase in lung volume may help to check for accuracy.
Introduction
In acute lung injury (ALI) and acute respiratory distress syndrome (ARDS), functional residual capacity (FRC) is markedly decreased as a result of numerous factors, including alveolar collapse, pulmonary edema with alveolar flooding, supine position, sedation-induced diaphragm inactivity, and cardiac enlargement [1-5]. Measuring FRC (or end-expiratory lung volume [EELV] when PEEP is applied) might help to measure the aerated lung available for ventilation and to better monitor the effects of ventilation strategies. Reproducible measurement techniques that can be used at the bedside are needed to minimize overdistention and to determine which patients may benefit from recruitment strategies. Repeated CT scans and gas-dilution techniques are two validated methods of lung-volume measurement but are so complex that their use has been confined to research settings. Recently, washout/washin techniques using oxygen [6,7] or nitrogen [8,9] have been made available in ICU ventilators, allowing bedside EELV measurement. A comparison of the nitrogen washout/washin EELV measurement with helium dilution or CT scan had shown good correlations in stable patients ventilated with low-PEEP levels [8]. The limitations of the nitrogen washout/washin technique for EELV measurement under other conditions, such as high FiO2 or high PEEP, have not been fully investigated [10].
PEEP-induced changes in lung volume (referred to as PEEP-volume) can also be assessed simply at the bedside by using passive spirometry. This accurate method requires a long expiration to zero end-expiratory pressure (ZEEP), where FRC is assumed to be reached. When considering the changes induced by two different levels of PEEP in a given patient, the difference in EELV (that is, ΔEELV = EELVhigh PEEP - EELVlow PEEP) should theoretically be similar to the difference in ΔPEEP-volume (PEEP-volumehigh PEEP - PEEP-volumelow PEEP), assuming that the FRC has not been modified by the PEEP changes (see Figure 1).
Figure 1. Schematic representation of the pressure-time diagram at the end of each epoch in
a single patient. Data at both PEEP levels studied are presented. Patm, atmospheric pressure; EELV, end-expiratory lung volume measured by using the nitrogen
technique; PEEP-volume, volume trapped by PEEP above the functional residual capacity,
measured by using a long exhalation to atmospheric pressure (zero end-expiratory pressure);
Vt, tidal volume; ΔEELV, EELVhigh PEEP - EELVlow PEEP ; ΔPEEP-volume, PEEP-volumehigh PEEP - PEEP-volumelow PEEP.
We designed a multicenter study with the primary objective of investigating the precision (reproducibility) of the nitrogen washout/washin technique for EELV measurement in patients with ALI/ARDS at two PEEP levels, including a high level, with a small variation in oxygen concentration (10%). Our secondary objective was to evaluate the accuracy of the technique by comparing PEEP-induced changes (Δ) in lung volume with the nitrogen technique or the PEEP-volume above the FRC measured with passive spirometry. As PEEP-volume is relatively easy to measure accurately with a calibrated pneumotachograph, it may therefore be considered a "gold standard." Because we expected possible discrepancies between the two techniques, we also compared the measured changes in lung volume (ΔEELV and ΔPEEP-volume) with the minimal predicted increase in lung volume, computed from static compliance (Cstat) at low PEEP and ΔPEEP. The minimal predicted increase in lung volume was considered the smallest-volume increase that can occur. We have also used this method to evaluate alveolar recruitment, as described elsewhere [11].
Material and methods
This was a multicenter study performed in five French medical intensive care units at the Henri Mondor University Hospital in Créteil, European Georges Pompidou University Hospital in Paris, Angers University Hospital in Angers, l'Archet 1 University Hospital in Nice, and Charles Nicolle University Hospital in Rouen. In compliance with French legislation, the institutional review board of the Henri Mondor University Hospital approved the protocol for all centers and waived the need for informed consent, as PEEP optimization was considered part of standard care. The patients or next of kin received information about the study.
Patients
Patients were enrolled if they met the standard criteria for acute lung injury (ALI) [12]: partial pressure of arterial oxygen over fraction of inspired oxygen (PaO2/FiO2) less than 300 mm Hg, bilateral pulmonary infiltrates on the chest radiograph, and no clinical evidence of left atrial hypertension. Most patients had ARDS, defined as PaO2/FiO2 less than 200 mm Hg. Exclusion criteria were age younger than 18 years, pregnancy, history of chronic obstructive pulmonary disease and/or lung surgery, and hemodynamic instability, defined as an increase in vasoactive drug (epinephrine, norepinephrine) dosages in the last 6 hours. All bedside anterior-posterior chest radiographs were reviewed by two independent observers (JJR and QL) according to CT Scan ARDS Study Group criteria to determine the pattern of aeration loss: lobar radiologic hyperattenuation predominating in the lower lobes (focal disease), diffuse radiologic hyperattenuation evenly distributed throughout the upper and lower lobes (white lungs), or patchy radiologic hyperattenuation involving the upper and lower lobes with persistent aeration of part of the upper lobes [13]. Patients with diffuse or patchy aeration loss were classified as having nonfocal disease [14].
Ventilation strategies
All patients received volume-assist control ventilation by using an Engström ICU ventilator (Version V4 and V5) with a CVOX module sensor (V4.5) General Electric, Madison (WI). This ventilator provides bedside EELV measurements by using the multibreath nitrogen-washout technique (MBNW) [8,15-18]. The oxygenation goal was achieved by adjusting FiO2, which was maintained constant during the study. Tidal volume was set at 6 ml/kg of predicted body weight. All patients received two PEEP levels, each for 45 minutes, in random order. PEEP levels were set as in the EXPRESS study [19]. In the minimal-distention strategy, PEEP and inspiratory Pplat were kept as low as possible while keeping arterial oxygen saturation at 88% to 92% or more. External PEEP was set to maintain total PEEP (the sum of external and intrinsic PEEP) between 5 and 9 cm H2O. In the optimized recruitment strategy, PEEP was adjusted based on Pplat and was kept as high as possible without increasing the inspiratory Pplat above 28 to 30 cm H2O. All patients were sedated. Neuromuscular blocking agents were administered only if deemed necessary by the clinician in charge.
Measurements
Lung volume and precision of measurements
At the end of each 45-minute period, blood was drawn for arterial blood gas measurement, and EELV was measured 3 times by using the MBNW technique to assess precision. This technique has been described elsewhere [9,16]. In brief, continuous measurement of end-tidal O2 and CO2 during a change in FiO2 (here, 10%) allows the calculation of nitrogen washout and washin and subsequently of the aerated lung volume. Two assumptions are made: heterogeneity in alveolar gas distribution is considered constant during the measurement procedure, and cellular metabolism and gas exchange between lung capillaries and alveoli are considered stable during the MBNW procedure. The mean of the washout and washin data is computed automatically if the difference between the two is less than 20% (cut-off determined by the manufacturer). Because FRC is a volume measured without PEEP (that is, at atmospheric pressure), we used the term end-expiratory lung volume (EELV) for the volume measured in our study. Three EELV measurements were performed at each PEEP level.
PEEP-volume (above FRC) by using passive spirometry
Prolonged exhalation (15 seconds) to the elastic equilibrium volume at ZEEP was performed, at the end of a 45-minute period, to standardize lung-volume history. Pressure and flow were recorded by using a dedicated computer linked to the ventilator (sample every 0.04 seconds), pressure, and flow curves were drawn off-line by using the software (Acknowledge 3.7.3) Goleta Ca. Volumes were measured by flow integration. PEEP-volume above FRC was obtained by subtracting the insufflated tidal volume from the flow-signal integration of this long exhalation. PEEP-volume was measured at the end of each of the two PEEP periods.
Measurement of compliance
Cstat of the respiratory system was computed by dividing tidal volume by Pplat (measured during an end-inspiratory pause (1 second)) minus total PEEP. Total PEEP was measured by using an expiratory pause (1 second).
A pressure-volume curve was obtained during low-flow inflation from the low PEEP level to 30 cm H2O to check that compliance (Clin) was linear or not decreasing within this range.
Minimal predicted increase in lung volume
The minimal predicted increase in lung volume [20] is the smallest possible lung-volume increase that can be induced by PEEP. It was computed from Cstat at low PEEP, as follows:
Minimal predicted increase in lung volume (milliliters) = CstatlowPEEP ·ΔPEEP
where ΔPEEP is the difference between high and low PEEP.
This minimal increase should be equal to (if no recruitment occurs) or smaller than (if alveolar recruitment occurs) ΔEELV and ΔPEEP-volume. We evaluated the slope of the pressure-volume curve during tidal inflation to check that compliance did not decrease over tidal inflation and, therefore, that the computed minimal increase was indeed the lowest possible increase that could occur.
Statistical analysis
All variables are described as median (interquartile range). Precision of the nitrogen technique results was assessed by calculating the coefficient of variation for the three pairs of washout/washin measurements. The coefficient of variation was calculated as the SD of the differences divided by the mean of all measurements. The Bland and Altman method [21] was used to evaluate reproducibility of the nitrogen technique and to evaluate agreement between ΔEELV and ΔPEEP-volume. The largest difference between the three EELV measurements at each PEEP level was plotted against the mean. Accuracy of the technique was assessed by comparing the changes in lung volume induced by the PEEP increase. ΔEELV was plotted against ΔPEEP-volume. Correlations were evaluated by using linear regression (r2). Paired values were compared by using the Wilcoxon test. The Fisher t test and Mann-Whitney U test were used when appropriate. Values of p smaller than 0.05 were considered significant.
Results
We studied 37 patients, of whom three were excluded from the analysis because of poor signal quality (two patients had unstable signals during PEEP-volume recording (spontaneous breathing), and one had greater than 20% differences between washout and washin values). Table 1 reports the main characteristics of the 34 patients included in the analysis, 32 with ARDS and two with ALI. Table 2 reports data on ventilation mechanics, ventilator settings, measured volumes, and calculated volumes at each PEEP level. Both PEEP strategies were well tolerated by all patients. No patients experienced any significant desaturation during the study measurements (EELV or PEEP-volume).
Table 1. Patient characteristics
Table 2. Arterial blood gas values and ventilation during the minimal-distention (low PEEP) and high-recruitment (high PEEP) periods
Precision of the nitrogen technique
The 34 patients had three pairs of EELV measurements at each PEEP level (that is, 204) (Figure 2). Of these measurements, six (2.9%), in six different patients (two at low PEEP and four at high PEEP) showed greater than 20% differences between washout and washin values and were therefore excluded. The coefficient of variability for the remaining measurements was 3.0% at low PEEP and 3.9% at high PEEP (p < 0.0001).
Figure 2. Largest difference (absolute values) between the three EELV values obtained in each
patient, as a function of EELV. The difference between the measured values was larger at high PEEP (p = 0.004). When expressed as a percentage of EELV, no difference was observed according
to PEEP level. Gray diamonds, low PEEP; solid circles, high PEEP; red diamond, mean
value at low PEEP; blue circle, mean value at high PEEP; vertical and horizontal bars,
the standard deviation.
The largest mean difference between the three pairs of EELV measurements was 81 ± 64 ml. The difference was larger at higher PEEP levels (53 ± 43 ml versus 108 ± 69 ml; p = 0.004) but was similar when expressed as a percentage of EELV (Figure 2). Mean FiO2 was 67 ± 17%; the highest FiO2 levels were not associated with greater EELV variability.
Comparison with PEEP-induced changes in lung volume and accuracy of the method
Minimal predicted increase in lung volume, ΔEELV, and ΔPEEP-volume are shown in Figure 3.
Figure 3. Median and interquartile range of minimal predicted increase in lung volume, ΔEELV,
and ΔPEEP-volume. Minimal predicted increase in lung volume, 330 (190 to 421) ml. ΔEELV, 402 (263
to 654) ml. ΔPEEP-volume, 585 (325 to 822) ml.
ΔEELV and ΔPEEP-volume were only modestly correlated with each other (Figure 4a) (ΔEELV = 62.4 + [0.73 ΔPEEP-volume]; r2 = 0.47). Bias between these two measuring methods was 97 ± 255 ml, with a 95% confidence interval for limits of agreement of -414 to 608 ml (Figure 5).
Figure 4. ΔEELV and ΔPEEP-volume correlation. (a) Correlation between ΔEELV and ΔPEEP-volume in all patients (r2 = 0.47). Straight line is correlation: ΔEELV = 62.4 + 0.7 ΔPEEP-volume. (b) Relation between the minimal predicted increase in lung volume and ΔEELV. Red dots: patients in whom measurement errors were detected; dashed line,
identity. (c) Correlation between ΔEELV and ΔPEEP-volume after exclusion of the four patients with
obvious ΔEELV measurement errors (r2 = 0.80). Straight line is correlation: ΔEELV = -42.1 + 1.0 ΔPEEP-volume.
The relation between the minimal predicted increase in lung volume and ΔEELV was dispersed (Figure 4b). In particular, four patients had ΔEELV values that were substantially lower than the minimal predicted increase in lung volume (red dots; Figure 4b), suggesting underestimation of the volume change by EELV measurement. All four patients received PEEP levels ≥ 16 cm H2O, compared with only five of the 30 remaining patients (p = 0.003), and three had focal aeration loss compared with only three (10%) of the 30 remaining patients (p = 0.01). FiO2 was high (80% ± 16%) in these patients but was not significantly higher than that in the other patients (p = 0.1). The cause of ARDS (pulmonary or extrapulmonary) was not associated with measurement discrepancies. The high PEEP values suggested possible occurrence of leaks that could invalidate the measurements. When we excluded these four patients whose ΔEELV values were lower than the predicted minimal increase in lung volume, the correlation between ΔEELV and ΔPEEP-volume became substantially stronger (r2 = 0.80; Figure 4c).
Discussion
The main results of this physiological study can be summarized as follows: (a) the MBNW technique at two PEEP levels provided reproducible EELV measurements with acceptable precision; and (b) compared with ΔPEEP-volume and the minimal predicted increase in lung volume due to PEEP, ΔEELV measured by using the nitrogen technique seemed accurate for measuring lung-volume variations induced by PEEP. In a few patients, however, the method could give erroneous results, especially in case of high pressures. Comparing with the minimal predicted increase in lung volume may help to detect these errors.
Nitrogen technique variability
The MBNW technique described by Olegard et al. [9] allows bedside EELV measurement by using small and safe FiO2 increases and decreases (± 10%). Precision was greater with larger FiO2 changes [8,16], because nitrogen changes were greater. The small (10%) FiO2 change used in our study may have contributed to the test-retest variability but was deemed safer for our hypoxemic patients. All measurements were performed at the steady state 45 minutes after a change in PEEP, and no other interventions likely to affect cardiac output were performed, the patients being considered stable. Fewer than 3% of the EELV measurements failed (greater than 20% difference between washout and washin). Because the technique used to measure EELV involves computing the mean of washin and washout values [9], we assessed test-retest variability without comparing washout with washin. The variability we found in patients with ALI or ARDS at each PEEP level was comparable to that reported by Olegard et al. [9], who studied chiefly postoperative patients. As with the helium-dilution technique, absolute variability of the nitrogen technique in our study increased with higher PEEP and higher EELV. However, variability relative to absolute lung volume did not differ for higher EELV values (Figure 2). The lower precision reported by the manufacturer for FiO2 > 70% was not replicated here, but the flawed measurements seemed to occur at higher FiO2 values.
PEEP-induced changes in lung volume
EELV values at low PEEP in our study were very low (less than 1,000 ml at low PEEP) and similar to values obtained previously by using CT scan [2,22] or helium dilution [23] in ARDS patients. PEEP-volume and EELV represent different volumes obtained with two totally independent methods. We thus compared lung-volume changes induced by PEEP. ΔEELV and ΔPEEP-volume; both evaluated the PEEP-induced lung volume increase. The correlation was good in some patients but poor in others (Figure 5). The variability of EELV values may have contributed to a poor correlation. We sought to detect obviously flawed data by using a third method. Katz et al. [20] demonstrated that the lung-volume increase induced by PEEP changes was larger than expected from the airway-pressure change and compliance at low PEEP, indicating progressive lung recruitment [11]. We therefore calculated the minimal predicted increase in lung volume induced by PEEP, which is easily derived from Cstat at low PEEP [20]. In addition, by tracing a pressure-volume curve over the tidal-volume range at low PEEP, we checked that compliance did not decrease significantly within this volume range, to ensure that no volume increase smaller than the calculated minimal increase could occur. This method might prove useful at the bedside to assess the lower ΔEELV limit. Any difference between ΔEELV and this minimal predicted increase in lung volume may be considered an estimate of alveolar recruitment [11]. ΔPEEP-volume may slightly underestimate the lung-volume change, because of the assumption that FRC is unchanged after exhalation from high or low PEEP (Figure 3). Yet recent data [24] suggest that FRC may increase after high PEEP compared with low-PEEP ventilation. We used a 15-second expiration to ZEEP to minimize this problem. Our analysis, made at two PEEP levels, shown elsewhere, suggested that FRC was stable for our measurements [11].
Obvious discrepancies occurred in four patients. All four patients had the highest set PEEP levels (> 16 cm H2O). Although not proven, it is very possible that microleaks due to high set PEEP may explain discrepancies by decreasing the EELVhigh PEEP measurement and therefore ΔEELV. The higher set FiO2 values in these four patients may have adversely affected measurement precision, although further studies are needed to evaluate this possibility. Patients with focal aeration loss are at higher risk of hyperinflation versus recruitment [25], and the lung-volume distribution due to PEEP depends closely on disparities in regional lung compliance [26]. Another hypothesis could be that EELV discrepancies in patients with higher PEEP and focal aeration loss may be related to differences in regional gas distribution. MBNW equilibration may be impaired by regional time-constant inequalities [27], and a higher dead space due to higher PEEP [28] and hyperinflation [29-31]. In clinical practice, we suggest comparing the increase in EELV with PEEP to the minimal predicted increase in lung volume to detect erroneous measurements.
Conclusions
The MBNW technique exhibits acceptable accuracy and precision for lung-volume measurement at different PEEP levels in patients with ARDS. Substantial underestimation of lung-volume changes may occur, at least in some patients, presumably in case of leaks due to high pressures, and additional measurements may be required to check this accuracy.
Key messages
• Nitrogen washin/washout technique exhibits acceptable accuracy and precision for lung-volume measurement at different PEEP levels and high FiO2 in patients with ARDS.
• Underestimation of lung-volume changes may occur in some patients presumably in case of leaks due to high pressures.
Abbreviations
ALI: acute lung injury: ARDS: acute respiratory distress syndrome; Clin: linear compliance; Cstat: static compliance; EELV: end-expiratory lung volume; FRC: functional residual capacity; MBNW: multibreath nitrogen washout; PEEP: positive end-expiratory pressure; PEEP-volume: trapped lung volume due to PEEP; Pplat: plateau pressure; Vt: tidal volume.
Competing interests
JD, LB, JCMR, AM, and their institution are involved in a patent with General Electric describing a method used to estimate alveolar recruitment. A grant was also received from General Electric for the conduct of the study. General Electric had no access to the data or to the content of the manuscript. All authors kept full control of the analysis of the data and the writing of the manuscript. JM and his team are currently doing research regarding FRC measurements, which is sponsored by a GE grant. NL, CS, GB, JLD, FDM, JJR, QL, and GB declare that they have no competing interests.
Authors' contributions
JD designed the study, and participated in data acquisition, statistical analysis, interpretation, and wrote the manuscript. NL participated in study design, data acquisition, statistical analysis, and manuscript editing. CS participated in data acquisition. GB participated in data acquisition. FDM participated in study design. AM, JCMR, JLD, and GB participated in study design and manuscript editing. JM participated in manuscript editing. JJR and QL participated in data analysis and manuscript editing. LB participated in study design, data analysis and interpretation, and manuscript writing. All authors read and approved the final manuscript.
Acknowledgements
General Electric provided the Engström ventilators for the study and a research grant, but had no access to the data, analysis, or interpretation.
References
-
Amato MB, Barbas CS, Medeiros DM, Magaldi RB, Schettino GP, Lorenzi-Filho G, Kairalla RA, Deheinzelin D, Munoz C, Oliveira R, Takagaki TY, Carvalho CR: Effect of a protective-ventilation strategy on mortality in the acute respiratory distress syndrome.
N Engl J Med 1998, 338:347-54. PubMed Abstract | Publisher Full Text
-
Gattinoni L, Caironi P, Cressoni M, Chiumello D, Ranieri VM, Quintel M, Russo S, Patroniti N, Cornejo R, Bugedo G: Lung recruitment in patients with the acute respiratory distress syndrome.
N Engl J Med 2006, 354:1775-86. PubMed Abstract | Publisher Full Text
-
Ranieri VM, Suter PM, Tortorella C, De Tullio R, Dayer JM, Brienza A, Bruno F, Slutsky AS: Effect of mechanical ventilation on inflammatory mediators in patients with acute respiratory distress syndrome: a randomized controlled trial.
JAMA 1999, 282:54-61. PubMed Abstract | Publisher Full Text
-
Rouby JJ, Constantin JM, Roberto De A, Girardi C, Zhang M, Lu Q: Mechanical ventilation in patients with acute respiratory distress syndrome.
Anesthesiology 2004, 101:228-34. PubMed Abstract | Publisher Full Text
-
Terragni PP, Rosboch G, Tealdi A, Corno E, Menaldo E, Davini O, Gandini G, Herrmann P, Mascia L, Quintel M, Slutsky AS, Gattinoni L, Ranieri VM: Tidal hyperinflation during low tidal volume ventilation in acute respiratory distress syndrome.
Am J Respir Crit Care Med 2007, 175:160-6. PubMed Abstract | Publisher Full Text
-
Patroniti N, Saini M, Zanella A, Weismann D, Isgro S, Bellani G, Foti G, Pesenti A: Measurement of end-expiratory lung volume by oxygen washin-washout in controlled and assisted mechanically ventilated patients.
Intensive Care Med 2008, 34:2235-40. PubMed Abstract | Publisher Full Text
-
Maisch S, Boehm SH, Weismann D, Reissmann H, Beckmann M, Fuellekrug B, Meyer A, Schulte Am Esch J: Determination of functional residual capacity by oxygen washin-washout: a validation study.
Intensive Care Med 2007, 33:912-16. PubMed Abstract | Publisher Full Text
-
Chiumello D, Cressoni M, Chierichetti M, Tallarini F, Botticelli M, Berto V, Mietto C, Gattinoni L: Nitrogen washout/washin, helium dilution and computed tomography in the assessment of end expiratory lung volume.
Crit Care 2008, 12:R150. PubMed Abstract | BioMed Central Full Text | PubMed Central Full Text
-
Olegard C, Sondergaard S, Houltz E, Lundin S, Stenqvist O: Estimation of functional residual capacity at the bedside using standard monitoring equipment: a modified nitrogen washout/washin technique requiring a small change of the inspired oxygen fraction.
Anesth Analg 2005, 101:206-12.
table of contents
PubMed Abstract | Publisher Full Text -
Brewer LM, Orr JA, Sherman MR, Fulcher EH, Markewitz BA: Measurement of functional residual capacity by modified multiple breath nitrogen washout for spontaneously breathing and mechanically ventilated patients.
Br J Anaesth 2011, 107:796-805. PubMed Abstract | Publisher Full Text
-
Dellamonica J, Lerolle N, Sargentini C, Beduneau G, Di Marco F, Mercat A, Richard JC, Diehl JL, Mancebo J, Rouby JJ, Lu Q, Bernardin G, Brochard L: PEEP-induced changes in lung volume in acute respiratory distress syndrome: two methods to estimate alveolar recruitment.
Intensive Care Med 2011, 37:1595-1604. PubMed Abstract | Publisher Full Text
-
Artigas A, Bernard GR, Carlet J, Dreyfuss D, Gattinoni L, Hudson L, Lamy M, Marini JJ, Matthay MA, Pinsky MR, Spragg R, Suter PM: The American-European Consensus Conference on ARDS, part 2. Ventilatory, pharmacologic, supportive therapy, study design strategies and issues related to recovery and remodeling.
Intensive Care Med 1998, 24:378-98. PubMed Abstract | Publisher Full Text
-
Malbouisson LM, Busch CJ, Puybasset L, Lu Q, Cluzel P, Rouby JJ: Role of the heart in the loss of aeration characterizing lower lobes in acute respiratory distress syndrome; CT Scan ARDS Study Group.
Am J Respir Crit Care Med 2000, 161:2005-12. PubMed Abstract | Publisher Full Text
-
Rouby JJ, Puybasset L, Nieszkowska A, Lu Q: Acute respiratory distress syndrome: lessons from computed tomography of the whole lung.
Crit Care Med 2003, 31:S285-95. PubMed Abstract | Publisher Full Text
-
Heinze H, Eichler W: Measurements of functional residual capacity during intensive care treatment: the technical aspects and its possible clinical applications.
Acta Anaesthesiol Scand 2009, 53:1121-30. PubMed Abstract | Publisher Full Text
-
Bikker IG, Scohy TV, Ad JJCB, Bakker J, Gommers D: Measurement of end-expiratory lung volume in intubated children without interruption of mechanical ventilation.
Intensive Care Med 2009, 35:1749-53. PubMed Abstract | Publisher Full Text | PubMed Central Full Text
-
Bikker IG, van Bommel J, Reis Miranda D, Bakker J, Gommers D: End-expiratory lung volume during mechanical ventilation: a comparison with reference values and the effect of positive end-expiratory pressure in intensive care unit patients with different lung conditions.
Crit Care 2008, 12:R145. PubMed Abstract | BioMed Central Full Text | PubMed Central Full Text
-
Scohy TV, Bikker IG, Hofland J, de Jong PL, Bogers AJ, Gommers D: Alveolar recruitment strategy and PEEP improve oxygenation, dynamic compliance of respiratory system and end-expiratory lung volume in pediatric patients undergoing cardiac surgery for congenital heart disease.
Paediatr Anaesth 2009, 19:1207-12. PubMed Abstract | Publisher Full Text
-
Mercat A, Richard JC, Vielle B, Jaber S, Osman D, Diehl JL, Lefrant JY, Prat G, Richecoeur J, Nieszkowska A, Gervais C, Baudot J, Bouadma L, Brochard L, Expiratory Pressure (Express) Study Group: Positive end-expiratory pressure setting in adults with acute lung injury and acute respiratory distress syndrome: a randomized controlled trial.
JAMA 2008, 299:646-55. PubMed Abstract | Publisher Full Text
-
Katz JA, Ozanne GM, Zinn SE, Fairley HB: Time course and mechanisms of lung-volume increase with PEEP in acute pulmonary failure.
Anesthesiology 1981, 54:9-16. PubMed Abstract | Publisher Full Text
-
Bland JM, Altman DG: Statistical methods for assessing agreement between two methods of clinical measurement.
Lancet 1986, 1:307-10. PubMed Abstract
-
Malbouisson LM, Muller JC, Constantin JM, Lu Q, Puybasset L, Rouby JJ: Computed tomography assessment of positive end-expiratory pressure-induced alveolar recruitment in patients with acute respiratory distress syndrome.
Am J Respir Crit Care Med 2001, 163:1444-50. PubMed Abstract | Publisher Full Text
-
Chiumello D, Carlesso E, Cadringher P, Caironi P, Valenza F, Polli F, Tallarini F, Cozzi P, Cressoni M, Colombo A, Marini JJ, Gattinoni L: Lung stress and strain during mechanical ventilation for acute respiratory distress syndrome.
Am J Respir Crit Care Med 2008, 178:346-55. PubMed Abstract | Publisher Full Text
-
Patroniti N, Bellani G, Cortinovis B, Foti G, Maggioni E, Manfio A, Pesenti A: Role of absolute lung volume to assess alveolar recruitment in acute respiratory distress syndrome patients.
Crit Care Med 2010, 38:1300-1307. PubMed Abstract | Publisher Full Text
-
Vieira SR, Puybasset L, Lu Q, Richecoeur J, Cluzel P, Coriat P, Rouby JJ: A scanographic assessment of pulmonary morphology in acute lung injury: significance of the lower inflection point detected on the lung pressure-volume curve.
Am J Respir Crit Care Med 1999, 159:1612-23. PubMed Abstract | Publisher Full Text
-
Puybasset L, Gusman P, Muller JC, Cluzel P, Coriat P, Rouby JJ: Regional distribution of gas and tissue in acute respiratory distress syndrome; III. Consequences for the effects of positive end-expiratory pressure; CT Scan ARDS Study Group, Adult Respiratory Distress Syndrome.
Intensive Care Med 2000, 26:1215-27. PubMed Abstract | Publisher Full Text
-
Pelosi P, Cereda M, Foti G, Giacomini M, Pesenti A: Alterations of lung and chest wall mechanics in patients with acute lung injury: effects of positive end-expiratory pressure.
Am J Respir Crit Care Med 1995, 152:531-7. PubMed Abstract | Publisher Full Text
-
Mekontso Dessap A, Charron C, Devaquet J, Aboab J, Jardin F, Brochard L, Vieillard-Baron A: Impact of acute hypercapnia and augmented positive end-expiratory pressure on right ventricle function in severe acute respiratory distress syndrome.
Intensive Care Med 2009, 35:1850-8. PubMed Abstract | Publisher Full Text | PubMed Central Full Text
-
Ventilation with lower tidal volumes as compared with traditional tidal volumes for acute lung injury and the acute respiratory distress syndrome; The Acute Respiratory Distress Syndrome Network
N Engl J Med 2000, 342:1301-8. PubMed Abstract | Publisher Full Text
-
Grasso S, Fanelli V, Cafarelli A, Anaclerio R, Amabile M, Ancona G, Fiore T: Effects of high versus low positive end-expiratory pressures in acute respiratory distress syndrome.
Am J Respir Crit Care Med 2005, 171:1002-8. PubMed Abstract | Publisher Full Text
-
Vieira SR, Puybasset L, Richecoeur J, Lu Q, Cluzel P, Gusman PB, Coriat P, Rouby JJ: A lung computed tomographic assessment of positive end-expiratory pressure-induced lung overdistension.
Am J Respir Crit Care Med 1998, 158:1571-7. PubMed Abstract | Publisher Full Text
-
Le Gall JR, Lemeshow S, Saulnier F: A new Simplified Acute Physiology Score (SAPS II) based on a European/North American multicenter study.
JAMA 1993, 270:2957-63. PubMed Abstract | Publisher Full Text
Figure 5.