 ResearchThe Bair Hugger patient warming system in prolonged vascular surgery: an infection risk?Joseph KC Huang1 , Elizabeth F Shah1, Narayanan Vinodkumar1, MA Hegarty2 and Robert A Greatorex3 1Surgical Registrar, Department of Surgery, Queen Elizabeth Hospital, King's Lynn, UK 2Consultant Pathologist, Department of Microbiology, Queen Elizabeth Hospital, King's Lynn, UK 3Consultant Surgeon, Department of Surgery, Queen Elizabeth Hospital, King's Lynn, UK author email corresponding author email
Critical Care 2003,
7:R13-R16doi:10.1186/cc1888 Abstract
Introduction
Use of the Bair Hugger forced-air patient warming system during prolonged abdominal vascular surgery may lead to increased bacterial contamination of the surgical field by mobilization of the patient's skin flora.
Methods
This possibility was studied by analyzing bacterial content in air and wound specimens collected during surgery in 16 patients undergoing abdominal vascular prosthetic graft insertion procedure, using the Bair Hugger patient warming system. The bacterial colony counts from the beginning and the end of surgery were compared, and the data analyzed using the Wilcoxon matched pairs test.
Results
The results showed not only that there was no increase in bacterial counts at the study sites, but also that there was a decrease (P < 0.01) in air bacterial content around the patient and in the operating theatre after prolonged use of the patient warmer. No wound or graft infections occurred.
Conclusion
The use of this warming system does not lead to increased bacterial contamination of the operating theatre atmosphere, and it is unlikely to affect the surgical field adversely. |