Minimal instructions improve the performance of laypersons in the use of semiautomatic and automatic external defibrillators

Stefan Beckers1*, Michael Fries1, Johannes Bickenbach1, Matthias Derwall2, Ralf Kuhlen3 and Rolf Rossaint4

Author Affiliations

1 Resident, Department of Anaesthesiology, University Hospital Aachen, Aachen, Germany

2 Medical Student, Department of Anaesthesiology, University Hospital Aachen, Aachen, Germany

3 Professor, Department of Anaesthesiology, University Hospital Aachen, Aachen, Germany

4 Professor and Chairman, Department of Anaesthesiology, University Hospital Aachen, Aachen, Germany

For all author emails, please log on.

Critical Care 2005, 9:R110-R116 doi:10.1186/cc3033


See related commentary http://ccforum.com/content/9/2/147

Published: 31 January 2005

Abstract

Introduction

There is evidence that use of automated external defibrillators (AEDs) by laypersons improves rates of survival from cardiac arrest, but there is no consensus on the optimal content and duration of training for this purpose. In this study we examined the use of semiautomatic or automatic AEDs by laypersons who had received no training (intuitive use) and the effects of minimal general theoretical instructions on their performance.

Methods

In a mock cardiac arrest scenario, 236 first year medical students who had not previously attended any preclinical courses were evaluated in their first study week, before and after receiving prespecified instructions (15 min) once. The primary end-point was the time to first shock for each time point; secondary end-points were correct electrode pad positioning, safety of the procedure and the subjective feelings of the students.

Results

The mean time to shock for both AED types was 81.2 ± 19.2 s (range 45–178 s). Correct pad placement was observed in 85.6% and adequate safety in 94.1%. The time to shock after instruction decreased significantly to 56.8 ± 9.9 s (range 35–95 s; P ≤ 0.01), with correct electrode placement in 92.8% and adequate safety in 97%. The students were significantly quicker at both evaluations using the semiautomatic device than with the automatic AED (first evaluation: 77.5 ± 20.5 s versus 85.2 ± 17 s, P ≤ 0.01; second evaluation: 55 ± 10.3 s versus 59.6 ± 9.6 s, P ≤ 0.01).

Conclusion

Untrained laypersons can use semiautomatic and automatic AEDs sufficiently quickly and without instruction. After one use and minimal instructions, improvements in practical performance were significant. All tested laypersons were able to deliver the first shock in under 1 min.

Keywords:
automated external defibrillator; cardiopulmonary resuscitation; defibrillation; layperson; intuitive